Oxygenation: Respiratory : Assessment of Oxygenation

Monitoring with Pulse Oximetry

Purpose

  1. Monitor arterial oxygen saturation (SaO2) noninvasively.
  2. Detect clinical hypoxemia promptly.
  3. Assess client's tolerance to tapering of oxygen therapy or activity.

Assessment

Equipment

Procedure

  1. Select appropriate type of sensor. A wide variety of sensors are available in sizes for neonates, infants, children, and adults. In addition, there are clip-on, adhesive, and disposable sensors. To select the appropriate sensor, consider the client's weight, activity level, if infection control is a concern, tape allergies, and anticipated duration of monitoring.
    Rationale: Proper sensor will increase accuracy of reading.
  2. Explain purpose of procedure to client and family.
    Rationale: Understanding the procedure increases compliance and prevents anxiety.
  3. Instruct client to breathe normally.
    Rationale: Consistent breathing prevents large fluctuations in minute ventilation and inaccurate reflections of SaO2 levels.
  4. Select appropriate site to place sensor. (Fig. 1) Avoid using lower extremities that may have compromised circulation, or extremities receiving infusions or other invasive monitoring. If client has poor tissue perfusion due to peripheral vascular disease or is receiving vasoconstrictor medications, a nasal sensor or forehead sensor may be considered.
    Rationale: Decreased circulation can falsely alter the SaO2 measurements from the pulse oximeter.
  5. Fig. 1: Select appropriate site with good tissue perfusion.

  6. Remove nail polish or acrylic nail from digit to be used.
    Rationale: Some nail polish (dark colors) and artificial nails can interfere with accurate measurements.
  7. Attach sensor probe (Fig. 2) and connect it to the pulse oximeter (Fig. 3). Make sure the photosensors are accurately aligned (Fig. 4).
    Rationale: Proper alignment is essential for accurate SaO2 measurement.
  8. Fig. 2: Attach sensor probe.

    Fig. 3: Connect sensor probe to pulse oximeter.

    Fig. 4: Check proper alignment of photosensors.

  9. Watch for pulse-sensing bar on face of oximeter to fluctuate with each pulsation and reflect pulse strength. Double-check machine pulsations with client's radial or apical pulse (Fig. 5).
    Rationale: A weak signal or missed pulsations will not produce an accurate measurement.
  10. Fig. 5: Select appropriate site with good tissue perfusion.

  11. If continuous pulse oximetry is desired, set the alarm limits on the monitor to reflect the high and low oxygen saturation and pulse rates. Ensure that the alarms are audible before leaving the client. Inspect the sensor site every 4 hours for tissue irritation or pressure from the sensor.
    Rationale: Ensure client safety by prompt detection of low critical oxygen saturation values or tissue irritation.
  12. Read saturation on monitor and document as appropriate with all relevant information on client's chart (Fig. 6). Report SaO2 less than 93% to physician.
    Rationale: Documentation provides healthcare team with baseline information and response to therapy. SaO2 of less than 93% usually indicates need for increased supplemental oxygen.
  13. Fig. 6: Read oxygen saturation from machine.

Sample Documentation
12/12/10 1400 O2 sat 93% on 2 L/m via cannula while resting in bed, O2 sat decreased to 88% when ambulated to the bathroom, RR increased to 26, MD notified and client returned to bed. O2 sat up to 93% when returned to bed and oxygen given at 2 L/m. Oxygen required during ambulation to prevent desaturation. S. Roberts, RN

Lifespan Considerations

Infant and Child Older Adult

Home Care Modifications

Collaboration and Delegation